To: | [email protected] |
---|---|
Subject: | Re: LF: Re: Re: FET RDS |
From: | Andy Talbot <[email protected]> |
Date: | Fri, 14 Jan 2011 11:22:08 +0000 |
Dkim-signature: | v=1; a=rsa-sha256; c=relaxed/relaxed; d=gmail.com; s=gamma; h=domainkey-signature:mime-version:in-reply-to:references:date :message-id:subject:from:to:content-type; bh=q17ePeLAcnENJZFrLwhtLI/mjcgeOey/9vEtwGETc64=; b=WHMWmuqVz05xpsm5WG+IPzqKbIRAbLNMkA2YaLhyL5f0DonsYHS5RhPLewQ50u+FOQ SCdDvzOX5Og/4SLupGkMOE/b5WNjHX+gz0NkTVzYUP9t6Gm+0hmTzBmb7EzhgfCLls29 wRJvolhjlJlfFkj/Tw0w3IHl+5bqLV2qDaCzw= |
Domainkey-signature: | a=rsa-sha1; c=nofws; d=gmail.com; s=gamma; h=mime-version:in-reply-to:references:date:message-id:subject:from:to :content-type; b=h+hlXCb9GigFW7KYymNlzwVwF6TYi4q3hC6j4ZXm5ddzLWM1UxxNBbHFqRqCuePJR/ BxJ9JjsUh9nkrhAAaod9D2pe2oicQtuwKuGKGMO347qSvlzV2gCXqDZ1Ug0fEFQ712NJ 4QAeCYRFPEePQ8DoYDQ/sv9rS1gLHDv3vsuxg= |
Domainkey-status: | good (testing) |
In-reply-to: | <002d01cbb3d3$ded8e9a0$0401a8c0@xphd97xgq27nyf> |
References: | <003801cbb33e$8b59e620$0401a8c0@xphd97xgq27nyf> <D9D2BC233E6248D095857032DE0B8974@JimPC> <004001cbb349$8b4ec190$0401a8c0@xphd97xgq27nyf> <[email protected]> <002d01cbb3d3$ded8e9a0$0401a8c0@xphd97xgq27nyf> |
Reply-to: | [email protected] |
Sender: | [email protected] |
Accurate LF Power measurement;
Its all about calibration, accuracy, attention to detail, and thinking about the sources of all errors.
If you think carefully about how to make measurements, even using amateur grade equipment you should be able to achieve decent accuracy. This is done by making most of the errors cancel out so you only have to rely on equipment that can be trusted, or verified elswhere
Here's how I make RF power measurements at LF/MF
Test equipment available:
DVM ; checked and compared with others and against IC voltage standard: Known accuracy < 0.2%
Clamp Ammeter - accuracy 1% - by checking against 2 other DMMs on current range ,but prone to a few tens of mA offset drift. But when measuring in the several-amp region this is within the 2%
Analogue 300MHz dual beam scope - which I regard as uncalibrated in absolute terms. However, it is linear over its full trace height , the attenuator volts/div is relatively accurate to 1% over the range 0.01 - 5V/div and the response is flat from DC to frequency of interest. All that has been checked out in the past Linearity and frequency response are all that matters
Bird 30dB 1kW attenuator, DC - 1GHz and 20dB coax attenuator (18GHz).
0- 50V DC variable PSU
Calibration Procedure:
Measure DC input resistance of Bird attenuator
Set scope for 50 ohm term input, fed from the 50dB of combined attenuators
Use DVM to set PSU to 50.0V DC output.and connect it to the input of the attenuator.
View scope trace of the resulting DC level. It should be "about" 0.158V, which corresponds to 50 Watts in and 50dB of attenuation
Adjust the scope "cal" control so the trace shows 1.58, or 3.16, or whatever, divisions.
(In actual fact the attenuation of the Bird is 29.54dB at DC, total 49.62dB - but that no longer matters - the scope trace is not calibrated)
I now know the scope, which has a known decent front panel attenuator is calibrated and errors in RF power attenuator and scope and are cancelled out.
The DC resistance of the Bird is 50.4 ohms - so only a potential 1% error there. That wouldn't have been taken out using the DC supply as reference
So I can now read voltage on the scope to an accuracy of about 2 - 3% - and know its correct.
PA Measurement
Apply RF from PA to attenuator input, viewing output sinewave the screen. Pout = (Vpk-pk)^2 / 400 * 100000.
Or a quick calculator Vp-p squared * 250
Don't need to asy much about the other half of the efficiency measurement. Even if the DVM weren't particulalrly accurate, the fact its the same as used for the DC voltage cal on the attenuator means any error cancels out.
So now the absolute sources of error are :
DVM calibration on one range - < 0.2%
DC Current clamp ammeter - 2%,
Scope readout resolution about 2% ish - depends if I squint hard enough!
Load resistance 1%
A perfectionist (or a contract-manager in my working days) would want errors added - so 6% or so potential uncertainty
A realist (or a contractor " " " " ) would use RSS addition of errors - so 3% realistic uncertainty
Which still - even after all this attention to detail, means a measured 90% efficiency could be 87 or 93%
My value of 80% was actually the lower limit of errors, so actual value could be in the region of 80 - 86%
Looking at the voltage across the FETs when on, especially at max Vdd, I can see wher eall teh loss is going - although the tank coil does run a bit warmer than I'd have hoped for. Its wound with 2mm Litz, and really needs thicker wire.
As an alternative - a 50 ohm resistor and schottky diode + decoupling C into a Hi-Z DVM make a good power detector for RF in the few watts region - like at the 30dB Bird output. PROVIDED it behaves as a peak hold detector, ie the decoupling cap does its job. A DC calibration is reasonably good for checking its ability to measure Vpeak (not pk-pk). The reading is no longer linear, but does lend itself to lookup tables, and an absoute cal against the DVM knowing the diode drop is mostly constant-ish
Whew.............!
Andy
On 14 January 2011 10:14, mal hamilton <[email protected]> wrote: Stefan |
<Prev in Thread] | Current Thread | [Next in Thread> |
---|---|---|
|
Previous by Date: | LF: Re: Mini-Whip Antenna, Petr Maly |
---|---|
Next by Date: | Re: LF: Mini-Whip Antenna, Tony |
Previous by Thread: | Re: LF: Re: Re: FET RDS, mal hamilton |
Next by Thread: | Re: LF: Re: Re: FET RDS, John Rabson |
Indexes: | [Date] [Thread] [Top] [All Lists] |